2 current measurements with one 5740A meter

Posted: Sun Apr 10, 2011 9:42 pm
by don0101
I have 2 AC currents to be measured. I intend to use a 100A:5A AC current shunt on each wire. Can I use a Double Throw switch to switch one 5740A meter between the 2 shunts without causing damage to the shunt that is not connected to the meter? Thanks.

Re: 2 current measurements with one 5740A meter

Posted: Mon Apr 11, 2011 2:44 am
by Tech_Marco
Yes you can. It won't damage the C.T. (current transformer) but both C.T. must be matched in size